I’ve been tracking a batch for 6 weeks. The vendor’s public COA lists 97.8% by HPLC-UV at 220 nm, one main peak at 11.2 min. I split the same vial and sent two aliquots to independent labs. Lab A: 93.2%, plus an extra peak at 8.4 min, 2.1% area.
Lab B: 96.9%, no extra peak, but their method runs 15 min with 0.1% TFA. Same sample, different columns and gradients. Is that 8.4 min peak real or a method artifact? What raw data would you ask both labs for—chromatograms, blanks, spiking?